FT660SC VOLTAGE IN - Monitored voltage is outside the range of 9.08–14.3 V
Check cable between sensor and controller.
2018-11-09 回复
FT660SC 5 VOLT - Monitored voltage is outside the range 4.5–5.5 V
If problem persists replace main circuit board
FT660SC TEMPERATURE - Sensor head internal temperature is higher than 50 °C.
Ensure that the environment is within the specifications.
FT660SC DARK READING - Dark reading detects too much light (more than 50 mNTU)
Perform 0 Electronics. Make sure head is on the body. Replace circuit board.
FT660SC LOW SIGNAL ERROR - Signal is too low (measuring less than 3.0 mNTU)
Make sure the head is seated properly. Check cleanliness of fiber optic detector.
FT660SC FLASH FAILURE - Datalog and event log will not work.
Replace main circuit board.
FT660SC LASER FAILURE - Laser failed to operate properly and is turned off.
Replace laser module.
FT660SC A/D FAILURE - A/D conversion failed
FT660SC Verification
Wet verification – perform verification with stablcalCVM verification – perform verification with calibration verification module
FT660SC Calibration
Calibrate quarterly with 2- 500mL bottles 800 mNTU standard