水智库平台

常见问题

  • FT660SC REFERENCE VOLTAGE

    Reference voltage is out of operating range of 2.41 and 2.59 V

    2018-11-09 回复

  • FT660SC PMT VOLTAGE PMT

    voltage is out of operating range of 11.77 and 15.94 V

    2018-11-09 回复

  • FT660SC OUTPUT MODE NOT NORMAL - Sensor is being calibrated.

    Complete calibration and/or verification.
    Release outputs

    2018-11-09 回复

  • FT660SC DATA LOG FULL/ EVENT LOG FULL - Only 30 minutes worth of space remains in the Data Log.

    Download log from the controller to prevent loss of data

    2018-11-09 回复

  • FT660SC LASER I

    Laser current is out of operating range of 25 and 75 mAmp

    2018-11-09 回复

  • FT660SC LASER V - Laser voltage is out of the operating range of 4.5 and 5.5 V

    Check connections on board

    2018-11-09 回复

  • FT660SC VOLTAGE IN - Monitored voltage is outside the range of 9.08–14.3 V

    Check cable between sensor and controller.

    2018-11-09 回复

  • FT660SC 5 VOLT - Monitored voltage is outside the range 4.5–5.5 V

    If problem persists replace main circuit board

    2018-11-09 回复

  • FT660SC TEMPERATURE - Sensor head internal temperature is higher than 50 °C.

    Ensure that the environment is within the specifications.

    2018-11-09 回复

  • FT660SC DARK READING - Dark reading detects too much light (more than 50 mNTU)

    Perform 0 Electronics.
    Make sure head is on the body.
    Replace circuit board.

    2018-11-09 回复