  • FT660SC Method of Operation

    2018-11-09 回复

  • FT660SC Body Assembly 60159-00

    2018-11-09 回复

  • FT660SC Head Assembly 60161-00

    2018-11-09 回复

  • FT660SC Low (

    Re-seat the head on the body.
    Laser failure
    PMT/Electronics failure

    2018-11-09 回复

  • FT660SC REFERENCE VOLTAGE

    Reference voltage is out of operating range of 2.41 and 2.59 V

    2018-11-09 回复

  • FT660SC PMT VOLTAGE PMT

    voltage is out of operating range of 11.77 and 15.94 V

    2018-11-09 回复

  • FT660SC OUTPUT MODE NOT NORMAL - Sensor is being calibrated.

    Complete calibration and/or verification.
    Release outputs

    2018-11-09 回复

  • FT660SC DATA LOG FULL/ EVENT LOG FULL - Only 30 minutes worth of space remains in the Data Log.

    Download log from the controller to prevent loss of data

    2018-11-09 回复

  • FT660SC LASER I

    Laser current is out of operating range of 25 and 75 mAmp

    2018-11-09 回复

  • FT660SC LASER V - Laser voltage is out of the operating range of 4.5 and 5.5 V

    Check connections on board

    2018-11-09 回复