CPU monitors 5 and 12 VDC, Lamp Voltage, and Lamp Current
2018-11-12 回复
SS7 Scheduled Maintenance
Instrument must be calibrated after any adjustment is made to the optical system!
2018-11-12 回复
SS7 Cleaning – Detector and light source
Detector Wipe lens using soft cloth and mild cleaning solution Light source Normal – Wipe shield assembly Extreme – Disassemble and clean lens and apertures (use great care if performing this procedure)
2018-11-12 回复
SS7 Light Source Assembly
2018-11-12 回复
SS7 Lamp replacement and alignment
Replace lamp annually Align light source Calibrate after lamp replacement
2018-11-12 回复
SS7 Calibration
Calibrate between 100 – 4000 NTU and verify lower level performance by comparison or by using StablCal standards. Not recommended: Calibration with standards lower than 100 NTU Calibration by comparison with lab instrument at low turbidity levels (<2 NTU)