  • SS7 Components

    2018-11-12 回复

  • SS7 Measurements interval

    Measurements are taken every 3 seconds

    2018-11-12 回复

  • SS7 Measure internal voltages

    CPU monitors 5 and 12 VDC, Lamp Voltage, and Lamp Current

    2018-11-12 回复

  • SS7 Scheduled Maintenance

    Instrument must be calibrated after any adjustment is made to the optical system!

    2018-11-12 回复

  • SS7 Cleaning – Detector and light source

    Detector
    Wipe lens using soft cloth and mild cleaning solution
    Light source
    Normal – Wipe shield assembly
    Extreme – Disassemble and clean lens and apertures (use great care if performing this procedure)

    2018-11-12 回复

  • SS7 Light Source Assembly

    2018-11-12 回复

  • SS7 Lamp replacement and alignment

    Replace lamp annually
    Align light source
    Calibrate after lamp replacement

    2018-11-12 回复

  • SS7 Calibration

    Calibrate between 100 – 4000 NTU and verify lower level performance by comparison or by using StablCal standards.
    Not recommended:
    Calibration with standards lower than 100 NTU
    Calibration by comparison with lab instrument at low turbidity levels (<2 NTU)

    2018-11-12 回复